    I have recently built a device for medical use, so electrical safety for the patient is of upmost importance.

    I currently use a medical grade switching Power Supply +/- 15VDC (IPD, model SRP-25-2005). An AC inlet, IEC 320, brings Line and Neutral power to a medical grade illuminated DPST (Multicomp, model MC3811-091), which in turn is connected to the power supply. Hope this makes sense!

    I would like to modify my current setup:

    1) Make the neutral continuous to the power supply without interruptions. It is currently connected to one pole of the rocker switch
    2) Use the +15VDC of the power supply to activate the light in the switch once it is "On"

    My question to you all:

    1) How can I do this?
    2) Do I have the right equipment (switch, power supply... )

    A simple diagram would be greatly appreciated!

    Thanks!

    Norm
    Stratmando's Avatar
    Stratmando Posts: 11,188, Reputation: 508
    Uber Member
     
    #2

    Oct 21, 2009, 03:34 PM

    Computers switch the neutral as well, I would have the neutrals tied together(not through switch)You don't want to use the Power switch, you will need to tie the new LED into the 15 Volts out. Dropping resistor will be needed.
    If you use an LED, I will supply the formula for the size resistor.
    KISS's Avatar
    KISS Posts: 12,510, Reputation: 839
    Uber Member
     
    #3

    Oct 21, 2009, 03:49 PM

    The datasheet is here: http://datasheet.octopart.com/MC3811...eet-537372.pdf

    But it's unclear.

    There probably should be some markings for line and load. If line is 240 V and connected to the LINE terminals, then the switch should illuminate when on. One might assume that it will work on 240 VAC as well.

    The LOAD terminals are not required to be connected.

    Your power supply is of a universal design, so I'm unsure what the line voltage markets are.

    I would suggest you contact Multicomp after verifying that the power cord is connected to the line terminals.

    Post any terminal designations from the bottom or side of the switch.
    KISS's Avatar
    KISS Posts: 12,510, Reputation: 839
    Uber Member
     
    #4

    Oct 21, 2009, 04:46 PM

    Take a look at this datasheet:

    http://www.rjselectronics.com/pdf/pad_PRS.pdf

    Note that you have to specify a lamp voltage because they are wired with a line and a load side.

    The RA9 has two small terminals for an LED, but is only SPST.
    The LED is usually current driven with about 15-20 mA. Thus to illuminate it, you would chose R near a standard value and select it to be less than or equal to R = 15/20 mA. A 1K resistor would give 15 mA.

    For R = 15/20 mA would yield a 750 ohm resistor is a standard value and could be used. Then the power disipation must be checked P = I^2*R = 0.02^2*750 = 0.3 W

    So, a 750 ohm 0.5 W resistor would work. LED's are polarity sensitive. Usually you would put the resistor in the positive side of the power supply.
    Stratmando's Avatar
    Stratmando Posts: 11,188, Reputation: 508
    Uber Member
     
    #5

    Oct 21, 2009, 05:57 PM

    You say you want an indicator on the output, How ever when Main Power is on, isn't Low Voltage present at the same time.
    An easy solution is a Bulb Across the Low Voltage output.
    For High Voltage, Check out the lumminated Appliance Rocker switch here, check out 4th from the bottom of the page:
    Rocker Switches
    Edit, just looked at kisses, rockers, The're fine.
    KISS's Avatar
    KISS Posts: 12,510, Reputation: 839
    Uber Member
     
    #6

    Oct 21, 2009, 06:09 PM

    But he wants medical approval. Not sure if the rockers I referenced have that.

Not your question? Ask your question View similar questions

 

Question Tools Search this Question
Search this Question:

Advanced Search

Add your answer here.


Check out some similar questions!

Power switch wiring [ 0 Answers ]

The pic enclosed is of a switch I need to wire as an on/off power switch. This switch is a 220/110 volt switch with a built in light. The way I have it wired in the pic I am just switching the power lead. Doing it that way the power goes off and on but no light shows I am using 110 volts...

Switch mode power supply [ 3 Answers ]

Please send a pitcher to me to make a switch mode power supply 13.8v dc at 10 amp

Wiring an illuminated rocker switch [ 6 Answers ]

Hi, I'd like to wire an illuminated rocker switch as an "on-off" switch for my AT power supply. The original switch had 4 wires to it, white, blue, brown, black. I had to bypass the old switch by looping the 4 wires in pairs, ergo,. hooked white to blue, then brown to black in order for the...

Wiring a ceiling fan to a switch without power [ 4 Answers ]

I just installed a new ceiling fan and want it to operate off an existing wall switch. The switch currently only operates a wall outlet in the same room. Apparently, the switch has no power, rather the power must be running to the wall outlet and then have a wire running to the switch. Is it...

3-Way Light Switch Illuminated Toggle Question [ 1 Answers ]

Hi: Here's my dilemma. I have an outdoor flood light that is outside of the garage and has both a 3-way switch in the garage and in the house. Since I put in those windowless garage doors, I can't see whether the flood light is on or off, and many a night someone leaves the light on all night....


View more questions Search