    Why Using the kcl means potassium chlorideinuv vis calibration
    Why Using the kcl means potassium chlorideinuv vis calibration
    Perito's Avatar
    Perito Posts: 3,139, Reputation: 150
    Ultra Member
     
    #2

    Dec 20, 2009, 11:08 AM

    Are you asking why is KCl used in UV/Vis calibration? I'm afraid I don't understand your question at all so let me take a guess at what you are asking.

    Stray light" is light from wavelengths other than the desired wavelength.

    A solution of (typically) 1.2% KCl is commonly used for stray light corrections in UV/Vis spectroscopy. The KCl provides ionic strength to the solution (hence, it changes the refractive index so as to be similar to comparable solutions), but it doesn't absorb in the UV or Visible region (except above its cut off wavelength). Therefore, light hitting the detector will be all stray light if you're selecting a wavelength where KCl does not absorb. KCl cuts off at 200 nM.
    InfoJunkie4Life's Avatar
    InfoJunkie4Life Posts: 1,409, Reputation: 81
    Ultra Member
     
    #3

    Dec 21, 2009, 07:19 AM

    I was curious about something similar... Does that mean it is basically a filter? Like in electronics, a low pass filter?
    Perito's Avatar
    Perito Posts: 3,139, Reputation: 150
    Ultra Member
     
    #4

    Dec 21, 2009, 12:05 PM

    What you're doing with the KCl is getting a reading of the stray radiation. You subtract that reading (it's a single number) from the single absorbance number you read when a sample is in the light path. Stray radiation causes readings to be too high. KCl is in there only to approximate the matrix that the sample would be in.

    This is a correction. It is not a filter of any kind. It doesn't have anything, that I can think of, in common with a low pass filter.

Not your question? Ask your question View similar questions

 

Question Tools Search this Question
Search this Question:

Advanced Search

Add your answer here.


Check out some similar questions!

Why 1.2 % KCL used for stray light calibration in UV Vis Spectrophotometer [ 4 Answers ]

Why 1.2 % KCL used for stray light calibration in UV Vis Spectrophotometer

Calibration of uv/vis spectrophotometer [ 8 Answers ]

Hi Please help me to find procedure of calibration of spectrophotometer by using reference materials (standard chemicals)

Uv vis calibration [ 1 Answers ]

Dear Sir, In calibration of uv vis spectrometer for resolution we use 0.02% toluene in hexane and calculate the ratio of absorbances at 266 and 269 nm which should be NLT 1.5. My question is 1) why we use haxane and toluene ? 2) why we should measure at 266 and why not other wavelengths?...

UV-VIS Spectrophotometer calibration using Pot.Dichromate [ 1 Answers ]

UV-VIS Spectrophotometer calibration using Pot.Dichromate

Calibration uv vis spectrophotometer [ 1 Answers ]

Calibration uv vis spectrophotometer


View more questions Search